Topic Title: can I calculate the calibration properties of an image post Acquisition?
Topic Summary: calculating image properties
Created On: 1/8/2016 5:58 AM
Status Post and Reply
Linear : Threading : Single : Branch
Topic Tools Topic Tools
View topic in raw text format. Print this topic.
 1/8/2016 5:58 AM
User is offline View Users Profile Print this message


Posts: 56
Joined: 5/16/2014

If you are going to be performing any form of Quantitative analysis on your image data then it would be more accurate to retrieve the calibration properties from the Acquisition system or re-acquire the data having first calibrated using the specific objective lens.

That said, the basic equation to derive the calibration properties (um/pixel) for your image is as follows:

image properties um/pixel = camera cell size/magnification*

e.g. for an image acquired using a Hamamatsu C9100-13 camera and a 40X objective lens (no other intermediate magnification) the calculation would be as follows:

cell size of the camera (taken from the Hamamatsu C9100-13 spec sheet) = 16um x 16um

magnification = 40

16/40 = 0.4 um/pixel

*magnification includes the Objective magnification used to acquire the data but may also include your C-mount magnification along with the magnification introduced by any other intermediate lens e.g. an Optovar.